Hi Alexis,
You are correct that the comments in the GetExtinction.m script state clearly the original spectra data are extinction coefficient, but Homer already does the
correction in GetExtinction line 534:
vLambdaHbOHb(:,2) = vLambdaHbOHb(:,2) * 2.303;
vLambdaHbOHb(:,3) = vLambdaHbOHb(:,3) * 2.303;
So the function does, as I said, output absorption coefficients, which is why all the latter functions use log base e.
Regards,
Rob
From: homer-users-bounces at nmr.mgh.harvard.edu [mailto:homer-users-bounces at nmr.mgh.harvard.edu]
On Behalf Of alexis machado
Sent: 03 July 2013 16:57
To: homer-users at nmr.mgh.harvard.edu
Subject: [Homer-users] FW: MBLL Missing ln(10)
Hi Robert thank you for you response
This is indeed very confusing
If i look in the header of GetExtinction.m it is sais clearly that
µa = (2.303) e x /(66,500) or in other terms µa = ln(10).e.[X] where [X] denotes molar concentration and e is the specific extinction coeff
Therefore i think that GetExtinction returns the specific extinction coefficients and not the specific absorption coefficient.
Moreover it seems confirmed in [1] See Important Note
Now, If i look in the HOmer's article [2] equation 2
ln (I/Io)=Dmua.L=(eps1.D[HbO]+eps2.D[HbR]) .L
Because It seems that HOMer use the specific molar extinction coefficient provided in [1] in the function GetExtinction.m and calculated using a base 10 logarithm. The Biomedical optic research laboratory website states that the ln(10) scale factor should
be considered for the calculation of Dmua
So , I think there is a MISSING ln(10) scale factor in equation 2 of this article and in any articles which use the extinction coefficients provided by reference [2] instead of specific absorption coefficients.
[1] http://www.medphys.ucl.ac.uk/research/borg/research/NIR_topics/spectra/spectra.htm
[2] P: HomER: a review of time-series analysis methods for near-infrared spectroscopy of the brain; Theodore J. Huppert, Solomon G. Diamond
Maria A. Franceschini and David A. Boas,
best regards
Alexis
From:
robert.cooper at ucl.ac.uk
To: alexis.machado at hotmail.fr
Subject: RE: [Homer-users] MBLL Missing ln(10)
Date: Wed, 3 Jul 2013 14:39:13 +0000
Hi Alexis,
This is a little confusing. I think the explanation is the difference between specific extinction coefficient (which is defined base 10, such that I = Io*10^(-Extinction*concentration))
and specific absorption coefficient (which is defined base e, such that I = Io*exp(absorption*concentration). I think that the homer function GetExtinctions, counter-intuitively, outputs specific absorption coefficients. The functions hmrIntensity2OD and
hmrOD2Conc are then all base e. Homer is therefore internally consistent, although the function GetExtinctions should perhaps be renamed.
Rob
Rob J Cooper PhD.
Research Associate
Biomedical Optics Research Laboratory
Malet Place Engineering Building, Rm 3.18
University College London
Gower Street
London WC1E 6BT
T: +44 (0)20 7679 0275
From:
homer-users-bounces at nmr.mgh.harvard.edu [mailto:homer-users-bounces at nmr.mgh.harvard.edu]
On Behalf Of alexis machado
Sent: 02 July 2013 18:36
To: homer-users at nmr.mgh.harvard.edu
Subject: [Homer-users] MBLL Missing ln(10)
Hi
I am a PhD student in NIRS.
I have a doubt on the definition of the Modified Beer Lambert Law as used in the sofware HOMER version1. Indeed HOmer uses a natural logarithm to define the absorbance ( also known as optical density) whereas the extinction coefficients values provided were
calculated with a log base 10 definition of the absorbance.
Let me explain much:
Let see the definition : REF HomER: a review of time-series analysis methods for near-infrared spectroscopy of the brain Theodore J. Huppert Solomon G. Diamond Maria A. Franceschini and David A. Boas you define the delta optical density or delta absorbance
as
equation: 1b
DOD=-ln(I/Io)=eps_1*D[HbO] +eps_2*D[HbR] *L % it is a natural logarithm definition
where D stands for variation from a baseline period, and OD the absorbance
where I is the detected fluence and Io the detected baseline fluence
where eps_1 is the molar extinction coeff of HbO and eps_2 the molar extinction coeff of HbR and L is the mean total optical pathlength
Because the beer lamber law states that I/Io=exp(-mua*L) equation 1.b implies that Dmua the absorption coeff variation= eps_1*D[HbO] +eps_2*D[HbR]
However according to the header explanation of the GetExtinctions.m function located in /HOMER/GeneralSubs/ the provided eps where measured using a different definition of the absorbance which is DOD=-log10(I/Io) . Indeed they say clearly that mua =ln(10)*(
eps_1*[HbO] +eps_2*[HbR])
Therefore, I think the equation 1.b should be corrected as in REF A review on continuous wave functional near-infrared spectroscopy and imaging instrumentation and methodology Scholkmann F, Kleiser S, Metz AJ, Zimmermann R, Mata Pavia J, Wolf U, Wolf M.
equation 1
DOD=-log10(I/Io) = eps_1*D[HbO] +eps_2*D[HbR] *L
Please correct me if i am wrong !!
best regards
Alexis
_______________________________________________
Homer-users mailing list
Homer-users at nmr.mgh.harvard.edu
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/homer-users
The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://mail.nmr.mgh.harvard.edu/pipermail/homer-users/attachments/20130703/0151de0f/attachment.html
More information about the Homer-users
mailing list