I really appreciate you checking this and getting to the bottom of it. I will make the changes to GetExtinction.m to clear up this confusion. 
Thx
David

Sent from my iPhone

On Jul 3, 2013, at 9:02 PM, alexis machado <alexis.machado at hotmail.fr> wrote:

> 
> Thanks
> 
> So all is OK
> 
> best regards
> Alexis
> 
> 
> 
> 
> 
> 
> Hi Alexis,
> 
>  
> 
> You are correct that the comments in the GetExtinction.m script state clearly the original spectra data are extinction coefficient, but Homer already does the correction in GetExtinction line 534:
> 
>  
> 
>         vLambdaHbOHb(:,2) = vLambdaHbOHb(:,2) * 2.303;
> 
>         vLambdaHbOHb(:,3) = vLambdaHbOHb(:,3) * 2.303;
> 
>  
> 
> So the function does, as I said, output absorption coefficients, which is why all the latter functions use log base e.
> 
>  
> 
> Regards,
> 
>  
> 
> Rob
> 
>  
> 
>  
> 
> From: homer-users-bounces at nmr.mgh.harvard.edu [mailto:homer-users-bounces at nmr.mgh.harvard.edu] On Behalf Of alexis machado
> Sent: 03 July 2013 16:57
> To: homer-users at nmr.mgh.harvard.edu
> Subject: [Homer-users] FW: MBLL Missing ln(10)
> 
>  
> 
>  
> 
>  
> 
> Hi Robert thank you for you response
> 
>  
> 
> This is indeed very confusing
> 
>  
> 
>  
> 
> If i look in the header of GetExtinction.m it is sais clearly that
> 
> µa = (2.303) e  x /(66,500)   or in other terms µa = ln(10).e.[X]  where [X] denotes molar concentration and e is the specific extinction coeff
> 
>  
> 
>  
> 
> Therefore i think that  GetExtinction returns the specific extinction coefficients and not the specific absorption coefficient.
> 
>  
> 
> Moreover it seems confirmed in [1] See Important Note
> 
>  
> 
>  
> 
>  
> 
> Now, If i look in the HOmer's article [2]  equation 2
> ln (I/Io)=Dmua.L=(eps1.D[HbO]+eps2.D[HbR]) .L  
> 
> Because It seems that HOMer use the  specific molar extinction coefficient provided in [1] in the function GetExtinction.m and calculated using a base 10 logarithm. The Biomedical optic research laboratory website states that the ln(10) scale factor should be considered for the calculation of Dmua
> So , I think there is a  MISSING ln(10) scale factor in equation 2 of this article and in any articles which use the extinction coefficients provided by reference [2] instead of specific absorption coefficients.
> 
>  
> 
>  
> 
>  
> 
> [1] http://www.medphys.ucl.ac.uk/research/borg/research/NIR_topics/spectra/spectra.htm
> 
>  
> 
>  
> 
> [2] P: HomER: a review of time-series analysis methods for near-infrared spectroscopy of the brain; Theodore J. Huppert, Solomon G. Diamond  Maria A. Franceschini and David A. Boas,
> 
>  
> 
> best regards
> 
> Alexis
> 
>  
> 
>  
> 
>  
> 
>  
> 
>  
> 
>  
> 
>  
> 
> From: robert.cooper at ucl.ac.uk
> To: alexis.machado at hotmail.fr
> Subject: RE: [Homer-users] MBLL Missing ln(10)
> Date: Wed, 3 Jul 2013 14:39:13 +0000
> 
> Hi Alexis,
> 
>  
> 
> This is  a little confusing.  I think the explanation is the difference between specific extinction coefficient (which is defined base 10, such that I = Io*10^(-Extinction*concentration)) and specific absorption coefficient (which is defined base e, such that I = Io*exp(absorption*concentration).  I think that the homer function GetExtinctions, counter-intuitively, outputs specific absorption coefficients.  The functions hmrIntensity2OD and hmrOD2Conc are then all base e.  Homer is therefore internally consistent, although the function GetExtinctions should perhaps be renamed.
> 
>  
> 
> Rob
> 
>  
> 
>  
> 
>  
> 
> Rob J Cooper PhD.
> 
> Research Associate
> 
> Biomedical Optics Research Laboratory
> 
> Malet Place Engineering Building, Rm 3.18
> 
> University College London
> 
> Gower Street
> 
> London WC1E 6BT
> 
> T: +44 (0)20 7679 0275
> 
>  
> 
>  
> 
> From: homer-users-bounces at nmr.mgh.harvard.edu [mailto:homer-users-bounces at nmr.mgh.harvard.edu] On Behalf Of alexis machado
> Sent: 02 July 2013 18:36
> To: homer-users at nmr.mgh.harvard.edu
> Subject: [Homer-users] MBLL Missing ln(10)
> 
>  
> 
> Hi
> I am a PhD student in NIRS.
> I have a doubt on the definition of the Modified Beer Lambert Law as used in the sofware HOMER version1. Indeed HOmer uses a natural logarithm to define the absorbance  ( also known as optical density) whereas the extinction coefficients values provided were calculated with a log base 10 definition of  the absorbance. 
> 
> Let me explain much: 
> 
> Let see the definition : REF  HomER: a review of time-series analysis methods for near-infrared  spectroscopy of the brain  Theodore J. Huppert  Solomon G. Diamond Maria A. Franceschini and David A. Boas you define the delta optical density or delta absorbance as 
> 
> equation: 1b
> DOD=-ln(I/Io)=eps_1*D[HbO] +eps_2*D[HbR] *L   % it is a natural logarithm definition
>  
> where D stands for variation from a baseline period, and OD the absorbance
> where I is the detected fluence and Io the detected baseline fluence
> where eps_1 is the molar extinction coeff of HbO and eps_2 the molar extinction coeff of HbR and L is the mean total optical pathlength
> 
> Because the beer lamber law states that I/Io=exp(-mua*L)  equation 1.b implies that  Dmua the absorption coeff  variation= eps_1*D[HbO] +eps_2*D[HbR] 
> 
> 
> However according to the  header explanation of the GetExtinctions.m function located in /HOMER/GeneralSubs/   the provided eps where measured using a different definition of the absorbance which is DOD=-log10(I/Io) . Indeed  they say clearly that  mua =ln(10)*( eps_1*[HbO] +eps_2*[HbR])
> 
> 
> Therefore, I think the equation 1.b should be corrected as in REF  A review on continuous wave functional near-infrared spectroscopy and imaging instrumentation and methodology Scholkmann F, Kleiser S, Metz AJ, Zimmermann R, Mata Pavia J, Wolf U, Wolf M.
> equation 1
> 
> DOD=-log10(I/Io) = eps_1*D[HbO] +eps_2*D[HbR] *L
> 
> Please correct me if i am wrong !!
> best regards
> Alexis
> 
> 
> _______________________________________________ Homer-users mailing list Homer-users at nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/homer-users The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
> _______________________________________________
> Homer-users mailing list
> Homer-users at nmr.mgh.harvard.edu
> https://mail.nmr.mgh.harvard.edu/mailman/listinfo/homer-users