Dear Experts,
For our project work, we particularly looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)? Could you please suggest to me a correct sequence for our project?
Regards Ramesh
External Email - Use Caution
Dear experts, Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence. For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)? Could you please suggest to me a correct sequence, if need to be changed?
Regards Ramesh
External Email - Use Caution
Hi,
I try to answer but Eugenio may also add comments. Several points that are important:
- 1.5 Tesla and 1x1x1 mm3 are at the limit of what the algorithm can manage.
- Post-aquisition reslicing to a higher resolution does not really increase the information content. Internally the algorithm works at a much higher resolution, anyway.
- If possible, switch to 3 Tesla.
- If possible and this is a prospective study, add a coronal T2 image and use the bi-channel option. See papers of Iglesias et al. on this.
- What you could also do on a 1.5 T is acquire the T1WI double and average it after coregistration, this increases the SNR.
- Generally, isometric resolutions are good due to the symmetry, of course.
- Also on 1.5 T, you may go bit below 1 mm with good head immobilization, and a longer scanning time. BRAVO in general is a good T1 option.
Best,
Philipp
________________________________ Von: freesurfer-bounces@nmr.mgh.harvard.edu freesurfer-bounces@nmr.mgh.harvard.edu im Auftrag von Ramesh Babu mgrameshbabu2013@gmail.com Gesendet: Freitag, 2. April 2021 09:29 An: Freesurfer support list; Umesh Rurdrapatna Betreff: Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution
Dear experts, Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence. For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)? Could you please suggest to me a correct sequence, if need to be changed?
Regards Ramesh
That all sounds right, thanks a lot Philipp!
Juan Eugenio Iglesias Senior research fellow CMIC (UCL), MGH (HMS) and CSAIL (MIT) http://www.jeiglesias.com
From: freesurfer-bounces@nmr.mgh.harvard.edu on behalf of "Saemann, Philipp" saemann@psych.mpg.de Reply-To: Freesurfer support list freesurfer@nmr.mgh.harvard.edu Date: Friday, April 2, 2021 at 03:51 To: Freesurfer support list freesurfer@nmr.mgh.harvard.edu, Umesh Rurdrapatna umeshrs@gmail.com Subject: Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution
Hi,
I try to answer but Eugenio may also add comments. Several points that are important:
- 1.5 Tesla and 1x1x1 mm3 are at the limit of what the algorithm can manage.
- Post-aquisition reslicing to a higher resolution does not really increase the information content. Internally the algorithm works at a much higher resolution, anyway.
- If possible, switch to 3 Tesla.
- If possible and this is a prospective study, add a coronal T2 image and use the bi-channel option. See papers of Iglesias et al. on this.
- What you could also do on a 1.5 T is acquire the T1WI double and average it after coregistration, this increases the SNR.
- Generally, isometric resolutions are good due to the symmetry, of course.
- Also on 1.5 T, you may go bit below 1 mm with good head immobilization, and a longer scanning time. BRAVO in general is a good T1 option.
Best,
Philipp
________________________________ Von: freesurfer-bounces@nmr.mgh.harvard.edu freesurfer-bounces@nmr.mgh.harvard.edu im Auftrag von Ramesh Babu mgrameshbabu2013@gmail.com Gesendet: Freitag, 2. April 2021 09:29 An: Freesurfer support list; Umesh Rurdrapatna Betreff: Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution Dear experts, Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence. For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)? Could you please suggest to me a correct sequence, if need to be changed?
Regards Ramesh
External Email - Use Caution
Dear Eugenio and Philipp,
Thank you very much for your suggestions. Could you please give more input on "acquire the T1WI double and average it after coregistration, this increases the SNR".
Should aquire twise the same sequence with 1x1x2 voxel dimension?
Thanks Ramesh
On Fri, 2 Apr, 2021, 9:00 PM Iglesias Gonzalez, Juan E., < JIGLESIASGONZALEZ@mgh.harvard.edu> wrote:
That all sounds right, thanks a lot Philipp!
Juan Eugenio Iglesias
Senior research fellow
CMIC (UCL), MGH (HMS) and CSAIL (MIT)
http://secure-web.cisco.com/1yAlNRge1To1jCEouu9JnBYJFnIJRo-86GCJ4EIWmAyj8pkk...
*From: *freesurfer-bounces@nmr.mgh.harvard.edu on behalf of "Saemann, Philipp" saemann@psych.mpg.de *Reply-To: *Freesurfer support list freesurfer@nmr.mgh.harvard.edu *Date: *Friday, April 2, 2021 at 03:51 *To: *Freesurfer support list freesurfer@nmr.mgh.harvard.edu, Umesh Rurdrapatna umeshrs@gmail.com *Subject: *Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution *Hi,
I try to answer but Eugenio may also add comments. Several points that are important:
- 1.5 Tesla and 1x1x1 mm3 are at the limit of what the algorithm can
manage.
- Post-aquisition reslicing to a higher resolution does not really
increase the information content. Internally the algorithm works at a much higher resolution, anyway.
If possible, switch to 3 Tesla.
If possible and this is a prospective study, add a coronal T2 image and
use the bi-channel option. See papers of Iglesias et al. on this.
- What you could also do on a 1.5 T is acquire the T1WI double and average
it after coregistration, this increases the SNR.
Generally, isometric resolutions are good due to the symmetry, of course.
Also on 1.5 T, you may go bit below 1 mm with good head immobilization,
and a longer scanning time. BRAVO in general is a good T1 option.
Best,
Philipp
*Von:* freesurfer-bounces@nmr.mgh.harvard.edu < freesurfer-bounces@nmr.mgh.harvard.edu> im Auftrag von Ramesh Babu < mgrameshbabu2013@gmail.com> *Gesendet:* Freitag, 2. April 2021 09:29 *An:* Freesurfer support list; Umesh Rurdrapatna *Betreff:* Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution *Dear experts,
Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence.
For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)?
Could you please suggest to me a correct sequence, if need to be changed?
Regards
Ramesh _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://secure-web.cisco.com/1IXus-BXNwSxxsWt0Hl4GNJkM1JdHvv3SmnEoV0r7QyM62L...
Twice the same image at 1x1x1, then use mri_robust_template to register and average them together
On 4/6/2021 8:10 AM, Ramesh Babu wrote:
External Email - Use Caution
Dear Eugenio and Philipp,
Thank you very much for your suggestions. Could you please give more input on "acquire the T1WI double and average it after coregistration, this increases the SNR".
Should aquire twise the same sequence with 1x1x2 voxel dimension?
Thanks Ramesh
On Fri, 2 Apr, 2021, 9:00 PM Iglesias Gonzalez, Juan E., <JIGLESIASGONZALEZ@mgh.harvard.edu mailto:JIGLESIASGONZALEZ@mgh.harvard.edu> wrote:
That all sounds right, thanks a lot Philipp! Juan Eugenio Iglesias Senior research fellow CMIC (UCL), MGH (HMS) and CSAIL (MIT) *MailScanner has detected a possible fraud attempt from "secure-web.cisco.com" claiming to be* http://www.jeiglesias.com <http://secure-web.cisco.com/1yAlNRge1To1jCEouu9JnBYJFnIJRo-86GCJ4EIWmAyj8pkkdBifGwyirgDc1KI-N6H7_dg4rBnLgE3CaKBeQuxL7GotNy9SzxKSD4OusIQrdGgcD_E0ObeWLPWYbYhFPgBDLEjcGPr-D8PT-k8bLD71quTve1q2fA_-MJY73fb-zPWX2QNDMT39_ceuRLWK8XGl18IGjSIXm6QeDK2erHCdQTO-rww0uRlSIph9mtjeVoEggFnM0sluSy0XceT-MPg9TJwCFg8RuNNq4J20pSw/http%3A%2F%2Fwww.jeiglesias.com> *From: *<freesurfer-bounces@nmr.mgh.harvard.edu <mailto:freesurfer-bounces@nmr.mgh.harvard.edu>> on behalf of "Saemann, Philipp" <saemann@psych.mpg.de <mailto:saemann@psych.mpg.de>> *Reply-To: *Freesurfer support list <freesurfer@nmr.mgh.harvard.edu <mailto:freesurfer@nmr.mgh.harvard.edu>> *Date: *Friday, April 2, 2021 at 03:51 *To: *Freesurfer support list <freesurfer@nmr.mgh.harvard.edu <mailto:freesurfer@nmr.mgh.harvard.edu>>, Umesh Rurdrapatna <umeshrs@gmail.com <mailto:umeshrs@gmail.com>> *Subject: *Re: [Freesurfer] 2mm thick image for hippocampal segmentation. * External Email - Use Caution * Hi, I try to answer but Eugenio may also add comments. Several points that are important: - 1.5 Tesla and 1x1x1 mm3 are at the limit of what the algorithm can manage. - Post-aquisition reslicing to a higher resolution does not really increase the information content. Internally the algorithm works at a much higher resolution, anyway. - If possible, switch to 3 Tesla. - If possible and this is a prospective study, add a coronal T2 image and use the bi-channel option. See papers of Iglesias et al. on this. - What you could also do on a 1.5 T is acquire the T1WI double and average it after coregistration, this increases the SNR. - Generally, isometric resolutions are good due to the symmetry, of course. - Also on 1.5 T, you may go bit below 1 mm with good head immobilization, and a longer scanning time. BRAVO in general is a good T1 option. Best, Philipp ------------------------------------------------------------------------ *Von:*freesurfer-bounces@nmr.mgh.harvard.edu <mailto:freesurfer-bounces@nmr.mgh.harvard.edu> <freesurfer-bounces@nmr.mgh.harvard.edu <mailto:freesurfer-bounces@nmr.mgh.harvard.edu>> im Auftrag von Ramesh Babu <mgrameshbabu2013@gmail.com <mailto:mgrameshbabu2013@gmail.com>> *Gesendet:* Freitag, 2. April 2021 09:29 *An:* Freesurfer support list; Umesh Rurdrapatna *Betreff:* Re: [Freesurfer] 2mm thick image for hippocampal segmentation. * External Email - Use Caution * Dear experts, Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence. For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension. Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)? Could you please suggest to me a correct sequence, if need to be changed? Regards Ramesh _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu <mailto:Freesurfer@nmr.mgh.harvard.edu> *MailScanner has detected a possible fraud attempt from "secure-web.cisco.com" claiming to be* https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer <https://secure-web.cisco.com/1IXus-BXNwSxxsWt0Hl4GNJkM1JdHvv3SmnEoV0r7QyM62LQaQ7GOor7RtRN0aQzfhFM7lc_9Uu5fcLfv9rDBmaLelquFxD0i5EeJz7dRB1fSkcrH9OXh7CcxyQUdtB4YnGDLHG1bIPRqDN88-MQD4DLOCFxhRwE-yMzpgybXDc7HTRizMdSWlMKBuA-Ut2s6ZjmZ1o1v6hOcCdFt73IOXO-n12alCMDKQF2h4U1gnWzOiaxIt45LuwnOsHIljxgM20xQqjqxBim8X_oH_yng5Q/https%3A%2F%2Fmail.nmr.mgh.harvard.edu%2Fmailman%2Flistinfo%2Ffreesurfer>
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
External Email - Use Caution
Dear Douglas
Thank you very much for your suggestions.
Regards Ramesh
On Tue, 6 Apr, 2021, 7:03 PM Douglas N. Greve, dgreve@mgh.harvard.edu wrote:
Twice the same image at 1x1x1, then use mri_robust_template to register and average them together
On 4/6/2021 8:10 AM, Ramesh Babu wrote:
External Email - Use CautionDear Eugenio and Philipp,
Thank you very much for your suggestions. Could you please give more input on "acquire the T1WI double and average it after coregistration, this increases the SNR".
Should aquire twise the same sequence with 1x1x2 voxel dimension?
Thanks Ramesh
On Fri, 2 Apr, 2021, 9:00 PM Iglesias Gonzalez, Juan E., < JIGLESIASGONZALEZ@mgh.harvard.edu> wrote:
That all sounds right, thanks a lot Philipp!
Juan Eugenio Iglesias
Senior research fellow
CMIC (UCL), MGH (HMS) and CSAIL (MIT)
*MailScanner has detected a possible fraud attempt from "secure-web.cisco.com" claiming to be* http://secure-web.cisco.com/1oAFYauOAeFEI_KJ-fkusgflgO_3xjRHgp7o7CJ3qp41OmNY... http://secure-web.cisco.com/1yAlNRge1To1jCEouu9JnBYJFnIJRo-86GCJ4EIWmAyj8pkkdBifGwyirgDc1KI-N6H7_dg4rBnLgE3CaKBeQuxL7GotNy9SzxKSD4OusIQrdGgcD_E0ObeWLPWYbYhFPgBDLEjcGPr-D8PT-k8bLD71quTve1q2fA_-MJY73fb-zPWX2QNDMT39_ceuRLWK8XGl18IGjSIXm6QeDK2erHCdQTO-rww0uRlSIph9mtjeVoEggFnM0sluSy0XceT-MPg9TJwCFg8RuNNq4J20pSw/http%3A%2F%2Fwww.jeiglesias.com
*From: *freesurfer-bounces@nmr.mgh.harvard.edu on behalf of "Saemann, Philipp" saemann@psych.mpg.de *Reply-To: *Freesurfer support list freesurfer@nmr.mgh.harvard.edu *Date: *Friday, April 2, 2021 at 03:51 *To: *Freesurfer support list freesurfer@nmr.mgh.harvard.edu, Umesh Rurdrapatna umeshrs@gmail.com *Subject: *Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution *Hi,
I try to answer but Eugenio may also add comments. Several points that are important:
- 1.5 Tesla and 1x1x1 mm3 are at the limit of what the algorithm can
manage.
- Post-aquisition reslicing to a higher resolution does not really
increase the information content. Internally the algorithm works at a much higher resolution, anyway.
If possible, switch to 3 Tesla.
If possible and this is a prospective study, add a coronal T2 image and
use the bi-channel option. See papers of Iglesias et al. on this.
- What you could also do on a 1.5 T is acquire the T1WI double and
average it after coregistration, this increases the SNR.
- Generally, isometric resolutions are good due to the symmetry, of
course.
- Also on 1.5 T, you may go bit below 1 mm with good head immobilization,
and a longer scanning time. BRAVO in general is a good T1 option.
Best,
Philipp
*Von:* freesurfer-bounces@nmr.mgh.harvard.edu < freesurfer-bounces@nmr.mgh.harvard.edu> im Auftrag von Ramesh Babu < mgrameshbabu2013@gmail.com> *Gesendet:* Freitag, 2. April 2021 09:29 *An:* Freesurfer support list; Umesh Rurdrapatna *Betreff:* Re: [Freesurfer] 2mm thick image for hippocampal segmentation.
External Email - Use Caution *Dear experts,
Sorry for reposting. I am expecting some response from experts, so that I can proceed with confidence.
For our project work, we are looking for hippocampal and its subfield volumes. Using GE 1.5T scanner, we tried a 3D BRAVO sequence with 1x1x1 resolution, 256x256x160 dimension with 8* flip angle which produced images with a lot of noise. Then we tried the same sequence with 1x1x2 resolution, and 256x256x80 FOV, which produced a clean image without noise. When we reformatted the images, it produces 0.5x0.5x1 resolution with 512x512x172 dimension.
Is this sequence good for hippocampal volume estimation and subfield segmentation (also for whole-brain VBM analysis)?
Could you please suggest to me a correct sequence, if need to be changed?
Regards
Ramesh _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu *MailScanner has detected a possible fraud attempt from "secure-web.cisco.com" claiming to be* https://secure-web.cisco.com/17suehtrGEpDtQ5EXLhVOfQrRIUslZ-oteK8Gfy-y1RsK9R... https://secure-web.cisco.com/1IXus-BXNwSxxsWt0Hl4GNJkM1JdHvv3SmnEoV0r7QyM62LQaQ7GOor7RtRN0aQzfhFM7lc_9Uu5fcLfv9rDBmaLelquFxD0i5EeJz7dRB1fSkcrH9OXh7CcxyQUdtB4YnGDLHG1bIPRqDN88-MQD4DLOCFxhRwE-yMzpgybXDc7HTRizMdSWlMKBuA-Ut2s6ZjmZ1o1v6hOcCdFt73IOXO-n12alCMDKQF2h4U1gnWzOiaxIt45LuwnOsHIljxgM20xQqjqxBim8X_oH_yng5Q/https%3A%2F%2Fmail.nmr.mgh.harvard.edu%2Fmailman%2Flistinfo%2Ffreesurfer