Dear FreeSurfer group,
Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
I would like to know the following:
a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard?
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk:
Dear FreeSurfer group,
Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
the schwartz atlas is built from 3T data, which has somewhat different contrast properties due to the field-strength dependence of T1
Hopefully somone else can answer b) as I don't remember.
cheers Bruce
On Tue, 17 May 2016, Kasper Jessen wrote:
I would like to know the following: a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard?
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk: Dear FreeSurfer group, Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk
Do you have a reference to the Schwartz atlas? I can't seem to find it.
Thanks in advance.
Best Kasper
2016-05-17 14:30 GMT+02:00 Bruce Fischl fischl@nmr.mgh.harvard.edu:
the schwartz atlas is built from 3T data, which has somewhat different contrast properties due to the field-strength dependence of T1
Hopefully somone else can answer b) as I don't remember.
cheers Bruce
On Tue, 17 May 2016, Kasper Jessen wrote:
I would like to know the following:
a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard?
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk: Dear FreeSurfer group, Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
no, I don't think there is one. One of the many things we never published....
sorry Bruce
On Wed, 18 May 2016, Kasper Jessen wrote:
Do you have a reference to the Schwartz atlas? I can't seem to find it. Thanks in advance.
Best Kasper
2016-05-17 14:30 GMT+02:00 Bruce Fischl fischl@nmr.mgh.harvard.edu: the schwartz atlas is built from 3T data, which has somewhat different contrast properties due to the field-strength dependence of T1
Hopefully somone else can answer b) as I don't remember. cheers Bruce On Tue, 17 May 2016, Kasper Jessen wrote: I would like to know the following: a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard? Best Kasper 2016-05-10 8:09 GMT+02:00 Kasper Jessen <kasper_j@cnsr.dk>: Dear FreeSurfer group, Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment. However, where can i find the documentation for the above? Best regards, Kasper -- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk
Hi Kasper,
On May 17, 2016, at 10:05 , Kasper Jessen kasper_j@cnsr.dk wrote:
I would like to know the following:
a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard?
If I recall correctly this is based on http://www.sciencedirect.com/science/article/pii/S1053811909006703: Improvement of brain segmentation accuracy by optimizing non-uniformity correction using N3
Type Journal Article Author Weili Zheng Author Michael W. L. Chee Author Vitali Zagorodnov URL http://www.sciencedirect.com/science/article/pii/S1053811909006703 Volume 48 Issue 1 Pages 73-83 Publication NeuroImage ISSN 1053-8119 Date October 15, 2009 Journal Abbr NeuroImage
So most likely it reduces the N# smoothing distance from the default 200 to 50
Here is the relevant part from recon-all (v5.3): if ($DoNuIntensityCor3T) then # params from Zheng, Chee, Zagorodnov 2009 NeuroImage paper # "Improvement of brain segmentation accuracy by optimizing # non-uniformity correction using N3" set NuIterations = 1; set cmd = ($cmd --proto-iters 1000 --distance 50) endif
A somewhat wrong, but intuitive rationale is with array coils the loops are smaller and hence the coil sensitivity profiles as well and these modulate the intensity inhomogeneity. But again that is probably wrong…
Best Regards Sebastian
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk: Dear FreeSurfer group,
Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
Thanks for the answers :-)
Sebastian: What do you mean with "A somewhat wrong, but intuitive rationale is with array coils the loops are smaller and hence the coil sensitivity profiles as well and these modulate the intensity inhomogeneity. But again that is probably wrong…"
What is it that is probably wrong?
Best regards, Kasper
2016-05-17 14:48 GMT+02:00 Moeller, Sebastian < sebastian.moeller1@rwth-aachen.de>:
Hi Kasper,
On May 17, 2016, at 10:05 , Kasper Jessen kasper_j@cnsr.dk wrote:
I would like to know the following:
a) what is the differences between the 3T Atlas (schwartz atlas) and the
standard atlas?
b) what does the 3T-specific NU intensity correction parameters do
differently than the standard?
If I recall correctly this is based onhttp://www.sciencedirect.com/science/article/pii/S1053811909006703: Improvement of brain segmentation accuracy by optimizing non-uniformity correction using N3
Type Journal Article Author Weili Zheng Author Michael W. L. Chee Author Vitali Zagorodnov URL http://www.sciencedirect.com/science/article/pii/S1053811909006703 Volume 48 Issue 1 Pages 73-83 Publication NeuroImage ISSN 1053-8119 Date October 15, 2009 Journal Abbr NeuroImage
So most likely it reduces the N# smoothing distance from the default 200 to 50
Here is the relevant part from recon-all (v5.3): if ($DoNuIntensityCor3T) then # params from Zheng, Chee, Zagorodnov 2009 NeuroImage paper # "Improvement of brain segmentation accuracy by optimizing # non-uniformity correction using N3" set NuIterations = 1; set cmd = ($cmd --proto-iters 1000 --distance 50) endif
A somewhat wrong, but intuitive rationale is with array coils the loops are smaller and hence the coil sensitivity profiles as well and these modulate the intensity inhomogeneity. But again that is probably wrong…
Best Regards Sebastian
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk: Dear FreeSurfer group,
Our subject sample has been obtained on a 3 Tesla MR scanner. For this
reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom
it is
addressed. If you believe this e-mail was sent to you in error and the
contains patient information, please contact the Partners Compliance
HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you
in error
but does not contain patient information, please contact the sender and
properly
dispose of the e-mail.
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
Hi Kasper,
On May 17, 2016, at 17:44 , Kasper Jessen kasper_j@cnsr.dk wrote:
Thanks for the answers :-)
Sebastian: What do you mean with "A somewhat wrong, but intuitive rationale is with array coils the loops are smaller and hence the coil sensitivity profiles as well and these modulate the intensity inhomogeneity. But again that is probably wrong…"
What is it that is probably wrong?
I would be amazed if that rationale was a truthful description of the full physical basis for the changes in inhomogeneities; and to be on the safe side (I never even discussed this with a physicist) I declare that it is incomplete and hence “wrong”. The last thing I want is to spread half-baked information as ground truth…
Best Regards Sebastian
Best regards, Kasper
2016-05-17 14:48 GMT+02:00 Moeller, Sebastian sebastian.moeller1@rwth-aachen.de: Hi Kasper,
On May 17, 2016, at 10:05 , Kasper Jessen kasper_j@cnsr.dk wrote:
I would like to know the following:
a) what is the differences between the 3T Atlas (schwartz atlas) and the standard atlas? b) what does the 3T-specific NU intensity correction parameters do differently than the standard?
If I recall correctly this is based on http://www.sciencedirect.com/science/article/pii/S1053811909006703:Improvement of brain segmentation accuracy by optimizing non-uniformity correction using N3
Type Journal Article Author Weili Zheng Author Michael W. L. Chee Author Vitali Zagorodnov URL http://www.sciencedirect.com/science/article/pii/S1053811909006703 Volume 48 Issue 1 Pages 73-83 Publication NeuroImage ISSN 1053-8119 Date October 15, 2009 Journal Abbr NeuroImage
So most likely it reduces the N# smoothing distance from the default 200 to 50
Here is the relevant part from recon-all (v5.3): if ($DoNuIntensityCor3T) then # params from Zheng, Chee, Zagorodnov 2009 NeuroImage paper # "Improvement of brain segmentation accuracy by optimizing # non-uniformity correction using N3" set NuIterations = 1; set cmd = ($cmd --proto-iters 1000 --distance 50) endif
A somewhat wrong, but intuitive rationale is with array coils the loops are smaller and hence the coil sensitivity profiles as well and these modulate the intensity inhomogeneity. But again that is probably wrong…
Best Regards Sebastian
Best Kasper
2016-05-10 8:09 GMT+02:00 Kasper Jessen kasper_j@cnsr.dk: Dear FreeSurfer group,
Our subject sample has been obtained on a 3 Tesla MR scanner. For this reason we used the -3T flag to enable the 3T-specific NU intensity correction parameters and usage of the special 3T atlas for Talairach alignment.
However, where can i find the documentation for the above?
Best regards, Kasper
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
-- Kasper Jessen, MD, PhD student Center for neuropsychiatric Schizophrenia Research (CNSR) Mental Health Centre Glostrup Nordre Ringvej 29-67 2600 Glostrup Tlf.: 24430136 E-mail: kasper_j@cnsr.dk eller kasper.jessen@regionh.dk _______________________________________________ Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
freesurfer@nmr.mgh.harvard.edu