Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just wondering
1) why it worked in the bottom image and it failed in the upper two (T1_sag_comparison.jpg) 2) it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
________________________________________ From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but we found out that the gray-white segmentation on half our subjects is bad and it leads to increased white matter volumes and decreased gray matter volumes. I have included a snap shot showing the difference in segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle =20 degrees. and use a snapshot one slice to show the good CNR. We are using Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
hard to say - they are low contrast and at least one has pretty bad movement artifacts. I would try setting the various gray and white matter thresholds to mri_segment and mris_make_surface in the expert opts stuff. If you can't figure out how to do that, send me the recon-all.log from a good subject I can put one together for you and send it back
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just wondering
- why it worked in the bottom image and it failed in the upper two (T1_sag_comparison.jpg)
- it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but we found out that the gray-white segmentation on half our subjects is bad and it leads to increased white matter volumes and decreased gray matter volumes. I have included a snap shot showing the difference in segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle =20 degrees. and use a snapshot one slice to show the good CNR. We are using Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
Thank you,
Attached is the scripts folder for one subject .
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 4:11 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just wondering
1) why it worked in the bottom image and it failed in the upper two (T1_sag_comparison.jpg) 2) it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
________________________________________ From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but we found out that the gray-white segmentation on half our subjects is bad and it leads to increased white matter volumes and decreased gray matter volumes. I have included a snap shot showing the difference in segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle =20 degrees. and use a snapshot one slice to show the good CNR. We are using Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
Hi Tilak
this one (attached) worked pretty well on one of the datasets you sent me. Run recon-all with -expert xopts.txt and the usual other args. I would be very cautious about interpreting results though. For example there is very little gray/white contrast in multiple regions such as perirolandic cortex, and any apparent difference/correlation in thicknesss may just be subject motion. Similarly, I can't distinguish gray from white in the cerebellum, but hippocampus and amygdala/basal ganglia might be ok
cheers Bruce
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Thank you,
Attached is the scripts folder for one subject .
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 4:11 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just wondering
- why it worked in the bottom image and it failed in the upper two (T1_sag_comparison.jpg)
- it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but we found out that the gray-white segmentation on half our subjects is bad and it leads to increased white matter volumes and decreased gray matter volumes. I have included a snap shot showing the difference in segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle =20 degrees. and use a snapshot one slice to show the good CNR. We are using Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
Thank you for recommending the expert options. I have run this on two subjects (rows) and you can see the post analysis results on the right (column) in the attached image . As you have mentioned it didnt work well in the cerebellum any other ways to improve in the inferior regions?
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 5:12 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Thank you,
Attached is the scripts folder for one subject .
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 4:11 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just wondering
1) why it worked in the bottom image and it failed in the upper two (T1_sag_comparison.jpg) 2) it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
________________________________________ From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but we found out that the gray-white segmentation on half our subjects is bad and it leads to increased white matter volumes and decreased gray matter volumes. I have included a snap shot showing the difference in segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/... WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle =20 degrees. and use a snapshot one slice to show the good CNR. We are using Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
It does look better. Can you send me some more images--in axial and coronal planes?
On Wed, Mar 30, 2016 at 4:26 PM, Ramtilak Gattu ramtilak.gattu@wayne.edu wrote:
Thank you for recommending the expert options. I have run this on two subjects (rows) and you can see the post analysis results on the right (column) in the attached image . As you have mentioned it didnt work well in the cerebellum any other ways to improve in the inferior regions?
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 5:12 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Thank you,
Attached is the scripts folder for one subject .
Thanks
Regards Tilak ________________________________________ From: Ramtilak Gattu Sent: Monday, March 28, 2016 4:11 PM To: freesurfer@nmr.mgh.harvard.edu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
Sorry, Actually just found out that the receiving coil is the 12 channel head coil and the transmitting coil is the body coil. I am attaching a snapshot showing the sag orientation of the three cases that I previously mentioned and uploaded for your review.
From these figures do you really think there is not enough contrast, just
wondering
- why it worked in the bottom image and it failed in the upper two
(T1_sag_comparison.jpg) 2) it failed in the left image and did good in the right image (T1_sag_comparison_2.jpg).
We have similar data in 25 of the subjects. Is there a way to salvage the existing data and improve our results we will certainly follow your input in changing the existing protocol and collect some data and run the analysis to see if it improves the results.
One more thought is if is there any other way around to increase the contrast by taking help from T2 image and run recon-all. Do you think it will help?
Thanks for you help .
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Monday, March 28, 2016 11:22 AM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon -all for better volume estmation
what was the transmit coil? I think I would either reduce the voxel sizes a bit or not use grappa. Probably go to 1.2 mm or so, which would be a 70% increase in SNR. That won't help the transmit bias field though if you have one. The 8 channel can also sometimes be a problem. The 32 channel coils work much better
On Mon, 28 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
The scans are done on 3T Siemens TRIO and 8 coil head channel with a
GRAPPA of 2.
Here is the screen shot of protocol used.
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
Thanks
Regards Tilak
From: Ramtilak Gattu Sent: Sunday, March 27, 2016 3:23 PM To: Bruce Fischl Subject: Re: [Freesurfer] improving white/grey segmentation in recon
-all for better volume estmation
Hey Bruce,
I have uploaded the files in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
I have also included the 001.mgz in there.
Thanks
Regards Tilak
From: Bruce Fischl fischl@nmr.mgh.harvard.edu Sent: Saturday, March 26, 2016 8:35 PM To: Ramtilak Gattu Subject: Re: [Freesurfer] improving white/grey segmentation in recon
-all for better volume estmation
can you send me the rawavg.mgz from the subjects's mri file instead of these dicoms (which I can't read) On Sat, 26 Mar 2016, Ramtilak Gattu wrote:
Hey Bruce,
Thanks for the quick reply.
I have uploaded the cases in
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
xWlpwBQnNAhG8%3d&docid=0a72431d82b4d4b9b924cccbdfb7f34f8
The voxel dimensions are 1x1x1 mm.
Thanks
Regards
Tilak
From: Ramtilak Gattu Sent: Friday, March 25, 2016 9:59 PM To: freesurfer@nmr.mgh.harvard.edu Cc: tilakgattu@hotmail.com Subject: improving white/grey segmentation in recon -all for better
volume estmation
Hello all,
We have run recon -all on some healthy control subjects T1 images but
we found out that the gray-white segmentation on half our subjects is
bad and it leads to increased white matter volumes and decreased gray
matter volumes. I have included a snap shot showing the difference in
segmentation from bad to good for 3 cases
https://waynestateprod-my.sharepoint.com/personal/ao3080_wayne_edu/_layouts/...
WPDE1klUw%3d&docid=0dd4546a8004f46578b7a2177b192806a
. The parameter in the sequence are TE=4.92 ms, TR=19 ms, Flip angle
=20 degrees. and use a snapshot one slice to show the good CNR. We are using
Freesurfer version 5.3.
Thanks
Regards
Tilak.
The information in this e-mail is intended only for the person to whom
it is
addressed. If you believe this e-mail was sent to you in error and the
contains patient information, please contact the Partners Compliance
HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you
in error
but does not contain patient information, please contact the sender and
properly
dispose of the e-mail.
Freesurfer mailing list Freesurfer@nmr.mgh.harvard.edu https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer
The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.
freesurfer@nmr.mgh.harvard.edu